OR015
   
 

Synchrony Versus ReSTOR: A Comparison Of Two Presbyopia Correcting Lenses.

1. Con Moshegov
2. Melissa Serna

Perfect Vision Laser Correction Centres, Sydney, Australia

Purpose: To compare the visual outcomes of eyes implanted with the dual optic accommodating Synchrony IOL (AMO) and the diffractive multifocal ReSTOR (Alcon) lenses.

Methods: Twenty eyes of 12 patients underwent phacoemulsification and IOL implantation. 10 eyes of 7 patients were implanted with the Synchrony lens and 10 eyes of 5 patients with the ReSTOR. Though the patients are closely matched for age and pre-operative refractive error, this was not a controlled or randomized study. Refractive outcome and unaided and corrected visual acuity for both distance and near were documented.

Results: The mean pre-operative spherical equivalent of eyes undergoing Synchrony implantation was +2.92D and that of the ReSTOR group was +2.10D. The following day the majority of Synchrony eyes had a visual acuity worse than 6/12; the majority with the ReSTOR had acuities better than 6/12. This reflected the transient myopic shift seen with the Synchrony in the first 6 weeks or so. The mean unaided VA at 1 month was 6/12 with the Synchrony and 6/7.5 with the ReSTOR. Nine of the ten eyes with the Synchrony corrected to 6/6 or better, 7 of the 10 eyes with the ReSTOR did. Four (40%) of Synchrony eyes and 10 (100%) of the ReSTOR eyes were able to read the J8 font or finer with best distance correction in place.

Conclusion: With emmetropia the ability of eyes to see at near was greater with the ReSTOR lens than the Synchrony. However, the best corrected acuity may be better with the Synchrony.


 
RANZCO