Chapter 46A
Video-Based Low-Vision Devices
JENNIE Y. KAGEYAMA, MELISSA W. CHUN, KENT W. SMALL and JOSEPH L. DEMER
Main Menu   Table Of Contents

Search

CLOSED-CIRCUIT TELEVISION
OPTIMAL IMAGE CHARACTERISTICS FOR READING
COMPUTER VIDEO DISPLAYS AND DIGITAL DEVICES
COSTS AND BENEFITS OF VIDEO VISUAL DEVICES
FUTURE HORIZONS FOR VIDEO-BASED LOW-VISION DEVICES
REFERENCES

Low vision devices, whether optical or video based, facilitate visual function and help allay limitations due to varying degrees of visual impairment.1,2 Visual impairment results in three forms of loss: loss of central acuity, which represents resolution; loss of sensitivity to contrast; and loss of visual field. Visual impairment defined strictly on the basis of central acuity can be categorized as moderate (20/70 to 20/160), severe (20/200 to 20/400), or profound (worse than 20/400). The most commonly used visual aids address the problem of reduced visual acuity and employ magnification to increase image size on the retina to circumvent the patient's limited visual resolution. However, optical magnifiers, such as loupes, telescopes, and convex lenses, achieve image enlargement at the expense of other image characteristics that may be crucial for patients with low vision, namely, image brightness, contrast, and useful visual field. All these factors contribute to the impediment for quality viewing.

In distinction to optical magnifiers, video-based image processing systems need not sacrifice other image properties to achieve magnification. Although beneficial for patients with different degrees of visual impairment, video-based devices are particularly useful for patients with severe to profound visual impairment when conventional low vision devices are determined to be inadequate. Fonda and colleagues3 found that the closed-circuit television (CCTV) was most appropriate when optical devices did not produce sufficient magnification. Video-based systems introduced major advantages for people with low vision. These advantages include higher levels of magnification without the light-losing characteristics of purely optical systems, the ability to manipulate the video output electronically to create an enhanced image with greater contrast, and the availability of a larger field of view than is possible with solely conventional solutions.4 In fact, the CCTV can electronically magnify up to 60× without significant optical aberration. In addition, binocularity is maintained and the posture to read the text on the screen is more natural than the often uncomfortable and specific positioning required by other high-power near-vision aids.5 The primary disadvantage of the CCTV is that the system is not portable. However, with the availability of new technology and the use of head-mounted systems, this limitation can also be overcome.

First suggested in 1959 by Potts and coworkers,6 the use of CCTV systems as a low vision device was technologically feasible and popularized in the early 1970s4 by groups such as Genensky at the Rand Corporation7 and Fonda and colleagues.3 Although CCTVs have always been quite useful for some low vision applications, recent advances in our understanding of visual physiology, as well as in image processing technology, have greatly expanded the potential value of video-based low vision devices. In this chapter, both CCTV technology and the emerging technology of video enhancement are presented, along with an introduction to the various video-based low vision devices currently available for the patient with low vision. Optimal reading characteristics also are discussed.

Back to Top
CLOSED-CIRCUIT TELEVISION
A CCTV system consists of a camera, an X-Y positioning platform under the camera for placement of the text to be scanned, a light source to illuminate the material, and a video display monitor.3,8 The CCTV user moves the camera or printed text systematically to scan from left to right and from top to bottom. The X-Y table was a significant innovation in the design of CCTV systems because it greatly enhanced the comfort of reading and writing. Magnification is achieved through use of a zoom lens on the camera, the large size of the video monitor screen, and reduction of the viewing distance of the patient. An effective magnification of even 140× is possible,3 although the average magnification most commonly employed is far less.9 Basically, two types of magnification are in operation with the use of the CCTV: projection magnification and relative distance magnification.5

Several approaches can be taken to calculate the magnification achieved by a CCTV.10 Total magnification (M) of the CCTV can be calculated from the product of projection magnification (X) and relative distance magnification (Y), M = (X)(Y). X is defined as the print size on the monitor divided by the actual print size of the text (maintaining the same units of measure, e.g., centimeters); Y is the reference distance (25 cm or 40 cm) divided by the working distance in centimeters. Of course, if the reference distance is not designated, it is virtually impossible to determine the total magnification of the system. A cleaner approach in assessing the exact magnification of a CCTV is to determine the equivalent power provided by such a system. This allows the practitioner to obtain the dioptric equivalent of the projection magnification–relative distance magnification interaction. Once the dioptric equivalent is obtained, it is then easy to compare across systems and select systems of equivalent powers for patient evaluation, such as handheld and stand magnifiers, microscopes, or telemicroscopes. Equivalent dioptric power, Deq, equals (X)(Z), where X is the print size on the monitor divided by the actual print size and Z is equal to the dioptric equivalent of the working distance, that is, the reciprocal of the working distance or viewing distance of the patient from the monitor (100/working distance in centimeters). Finally, one can obtain a quick estimate of the amount of projection magnification, disregarding working distance. A piece of graph paper is placed under the camera of the CCTV after the appropriate print size has been established on the screen for a patient's reading needs. Another piece of graph paper is then held up to the monitor, and the regular-size graph squares that fit into the length of one or two magnified graph squares on the screen are counted. By dividing the resultant smaller number into the larger number, the projection magnification can be grossly determined.10

Stand-mounted CCTV systems are commonly configured with television receivers, video monitors, or computer monitors. They provide high-contrast, gray-scale, and inverse video display modes with adjustments for levels of contrast and brightness. The characteristics of enhancement may need to vary with the spatial frequency content of the television picture and the progression of the user's ocular condition.5

Contrast polarities are shown in Figure 1, whereas different contrast intensities are shown in Figure 2. Video displays can achieve contrast of up to 96%.11 Because video image contrast is electronically controllable at levels below the maximum achievable level in a given device, all video displays can be easily adjusted to produce any desired level of contrast intensity less than the maximum. This means that display contrast can usually be set to a level greater than that of the original subject matter before the camera. More importantly, contrast polarity can be easily reversed by an inverting amplifier, usually incorporated in video cameras and in better-quality monitors. These features of contrast enhancement and polarity reversal are unique to video-based visual devices.

Fig. 1. Text in black-on-white contrast polarity on Optelec CCTV (top), and text in white-on-black contrast polarity (bottom). The white-on-black polarity is more readable by patients with low vision who have cloudy ocular media. (Photo courtesy of Optelec.)

Fig. 2. Words printed with different intensities of contrast: fortune at 90%, tearful at 30%, working at 10%, and visible at 3%. These values have been degraded somewhat by photographic reproduction. (Legge GE, Rubin GS, Luebker A: Psychophysics of reading: V. The role of contrast in normal vision. Vis Res 27:1165, 1987.)

Handheld systems typically consist of a camera and a control box with a cable to be connected to a TV or video monitor. The user holds the camera and moves it across a page, which often requires sufficient hand-eye coordination. Contrast intensity and polarity can also be controlled on these systems.

Studies have confirmed the overall effectiveness of CCTVs in improving the reading ability of individuals with all levels of visual impairment.12 Sloan,13 in her case report on CCTVs, identified the following advantages of such a system:

  1. Viewing the screen from a normal reading distance enables people with binocular vision to avoid excessive convergence demand.
  2. The zoom lens allows for rapid change in magnification.
  3. The reversal of contrast to a black background and a white foreground is often less fatiguing than a white background and a black foreground because of the decrease in glare.
  4. The CCTV can be used for handwriting.
  5. Higher levels of magnification are available than is possible with purely optical solutions.
  6. The use of an X-Y table is beneficial for persons with visual restrictions who have difficulty keeping their place when reading.

However, a common theme stressed in the literature and by practitioners is that the effective use of CCTVs is seldom achieved through mere trial and error; rather, both a comprehensive low vision examination and training in use of the system are critical to success.

The digital revolution has begun to change the design of CCTVs and will continue to do so at an accelerated pace. Images can be enhanced and modified in a multitude of ways, and systems can be endowed with more memory and storage capabilities. These devices may become more popular and useful for orientation and mobility purposes as well.4

Back to Top
OPTIMAL IMAGE CHARACTERISTICS FOR READING
In our complex society, the ability to read is crucial to everyday activity and to nearly all forms of employment. For this reason, reading performance is usually taken as a benchmark for visual function.

An inherent limitation of video technology is that the amount of information displayed on the screen is defined by the camera's field of view. Obviously, when characters are made larger, fewer can be simultaneously displayed on a screen. For both normally sighted readers and those with low vision, Legge and associates14,15 found that reading rates are generally maximized when four or more characters are simultaneously displayed. Display of a window of fewer than four characters slows reading, whereas presentation of more than four characters limits the available magnification for a given video screen size. There are exceptions to this rule, however, and certain patients may continue to exhibit increases in reading rate when a window of up to 20 characters is employed.16

The reading rate of normal subjects declines markedly for contrast below 10%, as letters appear to fade into the background (see Figure 2). Most readers with low vision exhibit a sensitivity to image contrast that resembles that of normal subjects reading faded text. In the majority of patients with low vision, the minimum image contrast required for best reading speed averages 3.9 times that required by normally sighted subjects.11 Contrast is such an important parameter in CCTV reading performance that most patients with low vision prefer to use all of the contrast available from a CCTV display. In fact, contrast sensitivity of individual patients has been found to be a better predictor of CCTV reading performance than Snellen acuity tested at distance.17 The ability to detect threshold variations in contrast, in both the normally sighted person and the patient with low vision, is generally a function of the spatial frequency content of the image. Stripes, called gratings, having sinusoidally varying brightness are usually used to test contrast sensitivity at defined frequencies. Spatial frequency is defined as the number of cycles of alternation of the bright and dim stripes per degree of visual angle. Normal subjects exhibit maximal sensitivity to contrast for spatial frequencies of 2 to 4 cycles per degree,18 although spatial frequencies of as much as 25 to 30 cycles per degree can be detected when contrast is very high. The upper limit of spatial frequency detection corresponds roughly to threshold visual acuity for high-contrast Snellen letters. The reduced acuity of patients with low vision is reflected in a lower limit of maximum spatial frequency detection, but in these patients, the frequency of peak contrast sensitivity is usually reduced as well.17 As noted previously, the absolute contrast sensitivity of patients with low vision is also reduced. This is seen in Figure 3, which plots contrast sensitivity (the reciprocal of contrast threshold) as it varies with spatial frequency. The shapes of contrast sensitivity functions of different patients may vary even though their Snellen acuities are the same. It is tempting to think that there may be a way to manipulate the characteristics of the image to compensate for deficiencies in the contrast sensitivity functions of individual patients in order to improve their reading performance. One simple way to do this is through magnification, which has the effect of reducing the spatial frequency content of the image by the amount of the magnification factor. Another method of image enhancement proposed by Lawton18 employs computerized spatial frequency compensation filters to enhance the midrange spatial frequencies of images to compensate for specific deficits in the contrast sensitivity functions of individual patients with low vision. Although this method reportedly improves reading speed and reduces the amount of magnification required for reading, these claims have not been subjected to clinical confirmation.

Fig. 3. Contrast sensitivity functions for a normally sighted subject (broken line) and several patients with low vision having various levels of acuity (solid lines). (Brown B: Reading performance in low vision patients: relation to contrast and contrast sensitivity. Am J Optom Physiol Opt 58:218, 1981; copyright by the American Academy of Optometry.)

Contrast polarity has no significant effect on reading performance of normally sighted persons.19 Black characters on a white background are as easily read by normal subjects as white characters on a black background. However, the overall brightness of these two types of presentations is markedly different. The page with the black background has a lower average brightness and thus less ability to produce glare than the white page. Rubin and Legge11 found that contrast polarity has a significant effect on reading rate in a minority of patients with low vision, namely, in patients with “cloudy” ocular media. Patients with cloudy media typically read white characters on a black background faster than black text on a white background. The authors attributed this difference to the scattering of background light by the ocular media to produce “veiling glare” that reduces effective image contrast at the retina. This hypothesis is further supported by experiments in which black characters are presented against a white background within a window surrounding the characters.20 Reading rate was reduced for larger windows of white background in a manner consistent with veiling glare. Patients with macular degeneration had no significant preference for contrast polarity, whereas patients with reduced central vision due to retinitis pigmentosa often preferred white characters on a black background.21 Because many of the patients with retinitis pigmentosa also had posterior subcapsular cataract, this preference may also be due to glare effects from cloudy media, although cystoid macular edema may also have played a role.

Among the many ways that CCTV systems enhance visual images is their ability to provide images in different colors. Under photopic conditions, the color of displayed characters has no significant effect on the reading rates of normal subjects, except for characters near the threshold of resolution.22 Even for patients with low vision, image color only occasionally has a significant effect on reading. When there is an effect, reading performance is most likely to be impaired for red and, to a lesser extent, blue characters.22 Reading performance is best with green or white characters. There may, however, be an overall subjective color or contrast polarity preference depending on the individual, regardless of ocular condition.

Studies have shown that varying screen colors on CCTV systems does not affect reading performance when screen luminance and contrast settings are equivalent. To investigate whether screen color is an important variable in the prescription of CCTV systems, Jacobs23 measured the visual performance of CCTV users on white, green, and amber screens. There was no obvious link between the low vision condition and preference or aversion to those colors. Luminance transmittance appeared to be a greater factor in reading ability and maximizing luminance contrast optimized reading performance. On the basis of these findings, it may be concluded that a low-vision clinic does not need more than one CCTV screen color for its visual assessments. When the screen color can be chosen, a white screen may be more versatile.

When CCTV low vision devices are employed, patients control the rate and pattern of scanning of the text according to their individual needs. Normal subjects generally read static, or nonmoving, text more rapidly than text that is constantly moving across the screen. For patients with low vision, however, reading rates for drifting text tend to be modestly higher than for static text. In a study performed by Goodrich and colleagues24 it was determined that reading durations with the CCTV are significantly greater than with optical aids. A possible advantage of the CCTV is that it is sufficiently flexible to compensate for substantial changes in the patient's visual efficiency.25

The optimal characteristics of a CCTV low vision aid can be summarized as follows: Magnification should be variable but should at least include an effective value of 30×. At the chosen level of magnification, screen size should be large enough to display at least four characters, preferably as many as ten characters. This implies that larger screen sizes will nearly always be advantageous over smaller ones, enabling higher magnification. Contrast intensity should be as high as technically possible, and ambient illumination or filters should be adjusted to reduce glare from the video screen so as to maintain display contrast. Contrast polarity using a black background assists in eliminating glare. White characters are probably optimal for readability and contrast, but green is also acceptable. The X-Y positioning capabilities of the CCTV system should permit the low vision reader to smoothly scan the text to be read. Finally, the mechanical arrangement of the monitor and camera should enable the patient to maintain a comfortable posture to facilitate prolonged reading.

Back to Top
COMPUTER VIDEO DISPLAYS AND DIGITAL DEVICES
The increasing pervasiveness of computers in society has been paralleled by a trend toward use of adaptive computers and computer-based video displays by patients with low vision.26 Computer-based video displays share all of the advantages of the CCTV devices discussed previously but in addition, present the opportunity to manipulate additional features.

Another opportunity for flexibility in character presentation with computer video displays is in the spacing between characters. This may be of value to patients with central scotomas, who are forced to read using peripheral retinal loci where crowding effects may be significant. Although character spacing has no effect on acuity, the time required for recognition of characters increases when multiple, closely spaced characters are presented to peripheral retina.27 The standard spacing between characters is 20% of character width for most printed text. In the peripheral visual field, this spacing has been found to increase the time required for recognition by 22% to 65% of the minimum recognition time for uncrowded characters. Since the optimum letter spacing may vary from patient to patient, adjustable character spacing would be a desirable feature of computer video displays. This feature is currently available with existing software packages for personal computers. Such software includes various magnification and contrast enhancing capabilities.

Most personal computer hardware and software include built-in options that permit a moderate (usually twofold) enlargement of characters on the screen. Other computers, such as the Macintosh, can display font sizes up to 70 point or more directly on the screen. (Conventional character displays are 10 or 12 point.) However, the larger characters may not be compatible with certain widely used software, such as spreadsheets, which become impractical when they are so extensive that they fill more than several times the area of the display screen. Modern laser and ink-jet printers generally have the desirable capability of producing very high contrast print with characters in a wide range of sizes useful for patients with low vision, although again this may not be compatible with certain applications, such as spreadsheets and graphics. Special software is available to enlarge computer video displays. There not only are programs that magnify text and icons, but also products that use voice-output programs that read the text to the user. Because new hardware and software products are constantly being released, it is best to contact local retail distributors and computer users groups for additional information.

A very effective, albeit costly, method of improving the readability of a computer video display is to increase the size, or diagonal measurement, of the monitor. Screen magnifiers that are optical magnifiers, can be placed directly over video display screens. These are inexpensive and versatile but provide magnification limited to only about 2×.28 Nonmagnifying filter screens can be placed over video displays to reduce glare and improve effective contrast. Of course, spectacle telemicroscopes, allowing close focus, or loupes may be used in conjunction with video displays for additional magnification benefits. Handheld magnifiers may also be used as adjuncts.

Previously, most video-based systems were mainly employed for reading. However, with advancements in technology as well as increased patient necessities and demands, several digital devices for viewing far and intermediate distances have been developed, along with more sophisticated reading devices to facilitate reading tasks.

A digital handheld monocular telescope has recently been introduced that combines high magnification at long, mid, and near ranges with contrast enhancement and CCTV configuration capabilities (Fig. 4). Shaped like a handheld video camera with ergonomic design, the device has the ability to perform in low ambient light, a benefit over conventional monocular telescopes, which tend to require more light, depending on the ocular condition (Betacom, Glendale, AZ).

Fig. 4. The Betacom VisAble Video Telescope provides magnification of distance images. (Photo courtesy of Betacom.)

A new version of the conventional stand magnifier has also been developed (Fig. 5). It consists of a 3× (7.6 D) illuminated light-emitting diode stand magnifier, which when attached to any television set, displays digital images on the screen with approximately 9× to 17× magnification. The construction of this device allows the option of using it as an optical stand magnifier or digital device (Eschenbach Optik of America, Inc., Ridgefield, CT).

Fig. 5. The Eschenbach Videolupe Stand Magnifier allows for flexibility as an optical or digital device. (Photo courtesy of Eschenbach Optik of America, Inc.)

Revolutionary camera designs enable people with low vision to perform daily tasks more effectively. To illustrate, there is now a compact camera that turns a full 180 degrees to magnify images in any position up to 30× with autofocus capability of distant, intermediate, and near ranges. The system can be made portable with a pair of lightweight glasses that display the magnified image within the battery-operated system. Another device, a portable handheld digital magnifier, enables the use of a similar pair of lightweight glasses with built-in monitor for reading text (Fig. 6) (Enhanced Vision Systems, Huntington Beach, CA).

Fig. 6. The Maxport incorporates display monitors in a pair of glasses for portability during reading tasks. (Photo courtesy of Charlie Martin and Enhanced Vision Systems.)

For distance viewing, a head-mounted camera system is available with an autofocus magnification range of up to 25×. The system has been updated to provide a more lightweight and compact design with a portable battery pack. The zoom feature and autofocus capability assists in discerning details with greater ease (Fig. 7) (Enhanced Vision Systems, Huntington Beach, CA). The head-mounted design enables hands-free use. The autofocus feature allows a wide range of viewing distances. Contrast enhancement and reversal permit individuals to select optimal images for different tasks. Automatic video gain control and maintenance of constant screen luminance are functions of particular benefit to patients with adaptation/glare difficulties. Finally, the system allows for binocular viewing and the possibility for a wider field of view. Disadvantages include weight, appearance, cost, complexity of operation, and potential problems of motion sickness and claustrophobia. When viewing through a magnifying system, retinal image motion is magnified, and the visual vestibulo-ocular reflex gain must be equal to the magnification to stabilize the retinal image. The systems are unsuitable for patients with serious hand/head tremors or patients who are physically unable to operate the controls or support the headset.29

Fig. 7. The Jordy 2 is a head-mounted device for viewing a variety of distance ranges. (Photo courtesy of Enhanced Vision Systems.)

A mini, desktop-style, portable video magnifier is available for reading and writing and weighs less than 3 pounds. Much like the conventional CCTV, this lightweight system has a magnification range of 4× to 50× and connects to most television sets and computer monitors (Freedom Vision, Mountain View, CA).

Back to Top
COSTS AND BENEFITS OF VIDEO VISUAL DEVICES
Although video-based visual devices are the most versatile and effective of the highly magnifying low vision devices, they are also the most costly. Particularly when coupled with personal computers, video-based devices may be uniquely effective in restoring employment or educational potential for the patient. The greater costs of these aids must be weighed against their benefits.8

The rate of success in the use of video-based visual devices was high even for the devices available in the mid 1970s9; there is some evidence that improved technology has provided additional benefit. Even with technically efficient devices, initial training in their proper use is essential, and additional training can provide further improvement in reading speed and duration.26,30 Although reading comprehension is probably not a good measure of the effectiveness of a visual device, Legge and coworkers31 demonstrated that the comprehension of low vision subjects reading with a video display is equivalent to that of normally sighted subjects when reading speed is the same relative to the subject's maximum rate.

Back to Top
FUTURE HORIZONS FOR VIDEO-BASED LOW-VISION DEVICES
The new generation of adaptive devices has incorporated the advancements of digital technology. Whereas older video-based low vision devices were heavy, bulky, and mechanically awkward, recent developments in solid state video technology are already providing substantial practical benefits to the patient with low vision. Besides the obvious advantages of reduced weight and bulk, these portable displays feature low power consumption. New portable video-based devices are constantly being researched, including a digital device so compact that it may be worn around the neck and carried in the palm of one's hand. Given the rapid progress in video display technology, further refinements in this area undoubtedly will occur and will probably be associated with decreasing prices. With technically innovative advancements, the ability to improve low vision devices has dramatically increased and the potential to improve a low vision patient's ability to function more efficiently in daily activities has increased as well.

Interfaces between CCTV displays and personal computers are also rapidly evolving. These may be of particular benefit to patients who use visual devices for reading printed text during concurrent work using a personal computer. A device is already available that splits the video screen to allow part of it to display the output of the CCTV camera and the remainder to display enlarged computer characters (Telesensory Corporation, Sunnyvale, CA; International Business Machines, Armonk, NY).

Enormous potential exists for the development of software for patients with low vision to enhance readability of computer video display screens and perform complex digital image processing using personal computers. It seems likely that the continued decline in the cost of computer technology as well as advancements in design, resulting in lighter, smaller products and display images of improved quality, will enable practitioners to more commonly prescribe these new low vision devices and will provide further options for prescribing. Video magnification technology has the potential to offer considerable improvements to a patient's visual functioning and his or her ability to perform activities of daily living, as compared with conventional low vision aids. Moreover, this technology may offer solutions to some of the issues raised by users of conventional devices. Clearly, the benefits of video-based and digital devices are significant, optimizing functionality and, undoubtedly, enhancing quality of life for low vision patients.

The authors have no financial interest in any of the low vision products mentioned in this chapter.

Back to Top
REFERENCES

1. Leat SJ, Fryer A, Rumney NJ: Outcome of low vision aid provision: the effectiveness of a low vision clinic. Optom Vis Sci 71:199, 1994

2. Humphrey R, Thompson G: Low vision aids—evaluation in a general eye department. Trans Ophthal Soc UK 105:296, 1986

3. Fonda GE, Thomas H, Schnur RN: Evaluation of closed-circuit television as an optical aid for the low vision patient. Trans Am Acad Ophthalmol Otolaryngol 79:468, 1975

4. Uslan MM, Shen R, Shragai Y: The evolution of video magnification technology. J Vis Impairment Blindness 90:465, 1996

5. Benjamin WJ, Borish IM: Borish's Clinical Refraction. Philadelphia, WB Saunders, 1998

6. Potts AM, Volk D, West SS: A television reader: as a subnormal vision aid. Am J Ophthalmol 47: 580, 1959

7. Genensky SM: Some comments on a closed circuit TV system for the visually handicapped. Am J Optom Physiol Opt 46:519, 1969

8. Goodrich GL, Quillman RD: CCTVs: choices and considerations. Vis Impairment Blindness 71:68, 1978

9. Goodrich GL, Apple LE, Frost A, et al: A preliminary report on experienced closed-circuit television users. Am J Optom Physiol Opt 53:7, 1976

10. Tucker LA, Brilliant RL: Low vision near systems II: Electronic magnification systems. In Brilliant RL (ed) Essentials of Low Vision Practice, pp 239-250. Woburn MA, Butterworth-Heinemann, 1999

11. Rubin GS, Legge GE: Psychophysics of reading: VI. The role of contrast in low vision. Vision Res 29:79, 1989

12. Goodrich GL: Applying video and microcomputer technology in a low vision setting. Ophthalmol Clin North Am 7:177, 1994

13. Sloan, LL: Evaluation of closed-circuit television magnifiers. Sight Sav Rev 44:123, 1975

14. Legge GE, Pelli DG, Rubin GS, Schleske MM: Psychophysics of reading: I. Normal vision. Vision Res 25:239, 1985

15. Legge GE, Rubin GS, Pelli DG, Schleske MM: Psychophysics of reading: II. Low vision. Vision Res 25:253, 1985

16. Lovie-Kitchin JE, Woo GC: Effect of magnification and field of view on reading speed using a CCTV. Ophthalmic Physiol Opt 8:139, 1988

17. Brown B: Reading performance in low vision patients: relation to contrast and contrast sensitivity. Am J Optom Physiol Opt 58:218, 1981

18. Lawton TB: Improved reading performance using individualized compensation filters for observers with losses in central fixation. Ophthalmology 96:115, 1989

19. Legge GE, Rubin GS, Luebker A: Psychophysics of reading: V. The role of contrast in normal vision. Vision Res 27:1165, 1987

20. Legge GE, Rubin GS, Schleske MM: Contrast polarity effects in low vision reading. In Woo GC (ed): Low Vision: Principles and Applications, pp 288–307. New York, Springer-Verlag, 1987

21. Ehrlich D: A comparative study in the use of closed-circuit television reading machines and optical aids by patients with retinitis pigmentosa and maculopathy. Ophthalmic Physiol Opt 7:293, 1987

22. Legge GE, Rubin GS: Psychophysics of reading: IV. Wavelengths effects in normal and low vision. J Opt Soc Am 3:40, 1986

23. Jacobs RJ: Screen color and reading performance on closed-circuit television. J Vis Impairment Blindness 84: 569, 1990

24. Goodrich GL, Mehr EB, Darling NC: Parameters in the use of CCTVs and optical aids. Am J Optom Vis Sci 57:881, 1980

25. Legge GE, Ross JA, Luebker A, LaMay JM: Psychophysics of reading: VIII. The Minnesota low-vision reading test. Optom Vis Sci 66:843, 1989

26. Steele RD, Goodrich GL, Hennies D, McKinley JA: Reading aid technology for blind persons: responses to a questionnaire of experienced users. Assist Technol 1:23, 1989

27. Rohrkaste F, Whittaker S, Higgins K: Optimum letter spacing for word recognition in central and eccentric fields. Invest Ophthalmol Vis Sci 30(suppl):397, 1989

28. Goodrich GL, McKinley JI: A guide to large print computer access. J Vis Rehabil 1:29, 1987