| Chapter 47 Phototherapeutic Keratectomy ROGER F. STEINERT and ANN Z. McCOLGIN Table Of Contents  | 
| The argon fluoride excimer laser, emitting ultrashort pulses at 193 nm, precisely etches the cornea with only submicron damage to the adjacent nonablated stroma. The freshly ablated surface supports rapid and stable re-epithelialization in advance of reformation of normal basal lamina complexes. Therefore, the cornea can be recontoured by the excimer laser in vivo. The dominant clinical interest in this laser has been for modification of the corneal optics, particularly for reshaping the cornea for the correction of refractive error. Superficial corneal opacities and irregularities also are treated by the laser as a therapeutic intervention termed phototherapeutic keratectomy (PTK). This chapter reviews the general principles, indications for, and clinical results of PTK. | 
| PRINCIPLES OF THE EXCIMER LASER | ||||||||||||||||
| The term excimer is a contraction of excited dimer. Excited dimers are molecules with little or no binding in the electron
ground state but a more closely bound upper energy state. Rare gas atoms
interact with a halogen molecule when stimulated to the upper state
by electrical discharge within the laser cavity. High-power ultraviolet (UV) radiation is emitted as the bound upper state rapidly
dissociates to the ground state.1–3 First developed in 1975,4 the excimer laser is used scientifically to perform research in physical
chemistry and pump dye lasers,5 and industrially to etch a variety of materials.6 Table 1 lists the UV wavelengths obtained from some common gas mixtures. Excimer
laser emission is inherently in short pulses, typically around 10 nsec, with
a repetition rate of 1 to 50 Hz.
 
 TABLE 1. Excimer laser emission
   
 
  | 
| PHOTOBIOLOGY OF EXCIMER LASER CORNEAL ABLATION | 
| Research in the early 1980s showed that excimer laser-generated UV light
can precisely etch a variety of polymers.7–9 Srinivasan and Leigh6 observed that the irradiated molecules are broken into small fragments
that are ejected into the surrounding atmosphere, a process called ablative
photodecomposition. Ablative photodecomposition of organic polymers
is attributed to the high absorption by the polymer of short UV radiation, confining
the effect to near the surface, and the high energy
of each UV photon. At 193 nm, a single UV photon has an energy of 6.4 eV, which
exceeds the covalent bond strength of many molecules. After
bond breakage occurs, intense local pressure in a confined volume ejects
the molecular fragments into the surrounding atmosphere.6,9–14 Direct bond breakage by a high-energy photon is a photochemical laser–material interaction. The relative contribution of photochemical and thermal mechanisms to the UV ablation of organic polymers remains controversial.15 At the shortest laser wavelengths, such as 193 nm, the high photon energy may result in a purely photochemical process of ablative photodecomposition. At longer wavelengths, absorbed photon energy leads to a local rise in temperature that causes etching through a photothermal process. At longer wavelengths, coagulation of protein may be expected adjacent to the ablation zone. Short laser pulses delivered at a low repetition rate help to limit local heating. Photochemical and photothermal effects of excimer laser wavelengths on the cornea result from absorption by solid elements. Water poorly absorbs wavelengths between 193 and 293 nm.1 The carbon–nitrogen peptide bond is believed to be the source of a strong protein absorption peak at approximately 190 nm.17 Most of the corneal amino acids are nonaromatic, and they account for collagen absorption that begins to rise at wavelengths of less than 260 nm and particularly less than 240 nm.18,19 Aromatic amino acids absorb more strongly at wavelengths of greater than 240 nm. A collagen absorption peak around 250 nm is attributed to a carbon–nitrogen enolized peptide linkage.17 The corneal glycosaminoglycans have similar absorption spectra, with peaks around 190 nm and minimal absorption at 248 nm.20 Nucleic acids are limited to the occasional keratocyte in the stroma but are more important chromophores in the epithelium, with strong absorption at both 248 and 193 nm.21,22Ascorbic acid, particularly concentrated in epithelial cells, has more absorption at 248 nm than at 193 nm.23 Mutagenesis and carcinogenesis are concerns with UV radiation. Almost all carcinogens have been shown to be mutagens.24 UV radiation-induced mutation parallels absorption by DNA.19 The low density of stromal keratocytes offers some protection against carcinogenesis resulting from stromal photoablation. In several studies, 193-nm irradiation did not cause mutagenic or carcinogenic cellular events. Nuss and colleagues examined unscheduled DNA synthesis, a measure of repair of pyrimidine dimers. Compared with a control incision made with a diamond knife, unscheduled DNA synthesis did not increase after 193-nm linear ablation; in contrast, a statistically significant increase occurred after 248-nm irradiation.25 Possible mechanisms for the decreased toxicity at 193 nm include absorption of that wavelength by protein surrounding the nucleus (a protein shield), lack of cytotoxicity of DNA photoproducts produced by 193-nm light, DNA damage that can readily be repaired by the cells, or such lethal damage that potentially mutagenic repair processes are not possible.26 In a skin model, DNA damage and subsequent cytotoxicity were least at 193 nm, intermediate at 308 nm, and greatest at 248 nm. Corneal irradiation at 193 nm also results in fluorescence between 295 and 425 nm.27 These emissions could be both mutagenic and cataractogenic; however, the highly attenuated energy of the fluorescence may not reach toxic levels.  | 
| LASER–CORNEA INTERACTION | 
In 1983, Trokel and coworkers first reported the precise and controlled etching of the cornea by an argon–fluorine (ArF) excimer laser.28 Puliafito and colleagues compared the histopathologic effects of linear cornea ablation at 193 and 248 nm.29 Both studies found excellent preservation of normal corneal stromal microstructure adjacent to the ablation zone at 193 nm (Fig. 1). The adjacent cornea remained optically clear. High-power transmission electron microscopy showed a submicron zone of electron density immediately adjacent to the ablation only. Kerr-Muir and associates first described a pseudomembrane that appears to seal cells and cellular nuclei transected by the laser beam.30 In contrast, at 248 nm, disorganization of the collagen microstructure extended into the adjacent stroma for more than 10 νm.29 The cornea immediately adjacent to the ablation showed a loss of transparency, which was indicative of thermal injury. Further, a study by Peyman and associates31 showed a significant coagulative effect from 308-nm excimer laser radiation with induced corneal necrosis, stromal opacification, and endothelial cell damage. Krueger and associates quantitated ablation rates. Etch depth per pulse plotted against radiant energy density generated sigmoid curves.32,33 The steep portion of the curve is approximately logarithmic; this finding corresponds to observations in simpler organic polymers. The inflection point of the curve is approximately 200 mJ/cm2 for 193 nm, which represents the energy density yielding the most efficient ablation. Puliafito and colleagues postulated that ablation at the inflection points may be clinically undesirable; at the inflection point small changes in laser output could alter the ablation rate dramatically.34 Ablation at higher energy densities, such as 400 to 600 mJ/cm2 per pulse, where the ablation rate has the flattest slope, might be preferable in situations where ablation rate is critical. Krueger and coworkers33 further observed that for wavelengths higher than 193 nm, the threshold for corneal ablation increases as the laser repetition rate decreases. At 193 nm, the threshold was constant despite varying repetition rates. This observation is consistent with the photochemical theory of excimer corneal ablation, in which the buildup of heat does not play an important role at 193 nm and thermal mechanisms of ablation are important for ablation at longer wavelengths. The effect of radiation density on corneal smoothness remains controversial. Fantes and Waring showed improved corneal surface uniformity with their scanning system at higher laser energy density, especially with 500 mJ/cm2 per pulse.35 In contrast, Berns and coworkers observed that such high levels of energy can be associated with ridges on the corneal surface, damage to the endothelium, and the formation of small pits on the stromal surface inside the linear ablation zones.36 Endothelial toxicity probably is minimal, although endothelial effects can be seen despite the nearly complete absorption of 193 nm photons within 1 microns of tissue. Dehm and coworkers found that in linear ablations, 193-nm incisions to 90% of corneal depth produced endothelial membrane disruption similar to that seen underlying diamond knife incisions at the same depth.37 Frank loss of endothelial cells underlying the incision was observed at 248 nm. Hanna and coworkers observed the appearance of electron-dense granules in the endothelium after surface ablation of rabbit corneas at 193 nm.38 These electron-dense granules migrated through Descemet's membrane over several weeks and then dissipated. The clinical significance of this observation is unclear. This phenomenon may be related to transient pressure injury to the endothelium. Zabel and colleagues measured pressures as high as 100 atm at the endothelium during ablation of the superficial stroma, but no frank disruption of the corneal endothelium occurred.39 The fragments ejected from the corneal surface are visible several hundred nanoseconds after the laser exposure. For 193-nm irradiation, the particles travel at an initial velocity of approximately 400 m/sec for the first 500 nsec, but then rapidly decelerate. The ejection of the particles ends within 5 to 15 νsec as the decelerating particles form a mushroom-shaped cloud (Fig. 2). 40 Increased exposure energy leads to increases in both the plume size and initial velocity of the ejected fragments. Analysis of the plume has identified numerous molecules that contain between 10 and 20 carbon atoms.41  | 
| THERAPEUTIC APPLICATIONS OF THE EXCIMER LASER | 
| The excimer laser has received much attention for its ability to reshape the corneal surface to correct ametropia. Although correction of postoperative refractive errors and anisometropia has been called therapeutic, PTK typically refers to the use of the excimer laser for the treatment of other corneal disease. This section discusses the application of excimer laser ablation for recurrent erosions, treatment of surface irregularities, removal of superficial opacities, and treatment of complications after photorefractive keratectomy (PRK) and laser in situ keratomileusis (LASIK).42 PTK has been used successfully in pediatric as well as adult diseases.43 | 
| RECURRENT CORNEAL EROSIONS | 
| Painful recurrent erosion syndrome, whether resulting from trauma or anterior
membrane dystrophy (Cogan's dystrophy or map-dot-fingerprint
dystrophy), results from abnormalities in the epithelial
basement membrane complex. Most of these patients respond to topical lubrication
therapy, hyperosmotics, bandage soft contact lenses, debridement
of the epithelium and basement membrane, or anterior stromal micropuncture. An
occasional patient continues to have painful recurrent
erosions despite all these measures. Superficial PTK can be curative in
that setting. The ablated anterior corneal surface appears to be highly
supportive of stable re-epithelialization. Although completely normal
reformation of the basal lamina complex, including normal density
of hemidesmosomes and anchoring fibrils, may take months to years, most
investigators have been impressed with the rapid and stable re-epithelialization
that occurs after ablation, with absence of punctate keratitis, staining
defects, or symptoms of recurrent erosion. The objective of PTK in the treatment of recurrent erosions is simply to remove enough of the superficial Bowman's layer to permit formation of a new basement membrane with adhesion structures; therefore, it is the most superficial of the PTK procedures. The usual technique is to debride the loose epithelium manually. In most cases, even if the area of clinical recurrent erosion is localized, the surgeon finds that the epithelium is poorly adherent over most or all of the corneal surface. With the use of a large spot size, such as 6.5 mm, the surgeon ablates the central superficial corneal surface. Depending on the laser, varying techniques are used to expose the peripheral cornea outside the central zone. For example, with the VISX laser, a smaller spot is manually scanned around the periphery by moving the patient's head. With some smaller spot scanning lasers, the periphery can be treated by programming the scan to include these areas. The amount of tissue to be ablated to maximize the effect with minimal refractive change has not been elucidated. Some investigators remove only 2 to 6 microns,44–47 but others have found that removal of 10 microns results in fewer recurrences.48 After the laser ablation, a drop of topical antibiotic, steroid, and sometimes nonsteroidal agent is applied, followed by placement of a bandage contact lens. Close follow-up is required until epithelialization is complete, at which time the contact lens is removed. Patients typically are instructed to continue aggressive lubrication, including nightly application of lubricating ointment, for several months after the procedure. The results of clinical studies of PTK for recurrent erosions have been encouraging. Morad and coworkers42 treated 23 eyes for recurrent or persistent corneal erosion with PTK by ablating 2 to 6 νm of tissue after epithelial removal. Eighty-three percent of patients remained free of symptoms during the 12- to 60-month follow-up, and none experienced a significant refractive change. Cavanaugh and colleagues treated 36 eyes for recurrent erosion in the setting of anterior basement membrane dystrophy and found that 86.6% had no recurrence of symptoms after 12 months of follow-up.49 The average change in refraction was +0.72 ± c1.26 diopters (D), and a positive correlation was found between the number of pulses applied and induced refractive change. Jain and Austin43 reported that of 77 eyes treated with PTK for recurrent erosions, 69% were free of all symptoms after 6 to 55 months of follow-up and 100% were free of acute episodes after one or two treatments. It also was found that patients treated for recurrent erosions resulting from trauma had a better outcome than those whose symptoms were idiopathic or caused by a basement membrane dystrophy. Several other studies have shown a 60% to 100%44,45,50–56 cure rate with one or more PTK treatments for recurrent erosions. Quantitative studies indicate that the fundamental health of the corneal epithelium may be improved after PTK.57 More studies are necessary to determine the optimal amount of tissue to be removed to minimize both recurrences and significant refractive changes. Sridhar, Rapuano, and coworkers found that diamond burr polishing was at least as effective although simpler and less costly than PTK for recurrent erosions.58 In a similar mode, PTK has successfully reduced painful bullae in eyes with pseudophakic bullous keratopathy.59,60  | 
| IRREGULAR SURFACE | 
| Surface irregularities can severely disrupt the optical performance of
the anterior cornea. In addition to diminished visual acuity, patients
may complain of monocular diplopia, glare, halos, or other optical aberrations. Common
causes of surface irregularities include anterior corneal
dystrophies (including basement membrane dystrophy, Reis-Bucklers' dystrophy, Schnyder crystalline corneal dystrophy, and relatively
anterior granular and lattice stromal dystrophies), elevated
corneal scars (including apical scars associated with keratoconus), and
degenerations such as Salzmann nodule formation and band
keratopathy.61–63 Irregular astigmatism also can occur after refractive surgery, which is
discussed in further detail in the following. Many of these sources of surface irregularity also are associated with corneal opacities. In some cases, smoothing the optical surface alone significantly improves the patient's visual function without the hyperopic shift and potential scarring that can be associated with the deeper ablations often necessary for removal of opacities. Hard contact lens refraction can be helpful in determining what component of a patient's visual complaints is secondary to surface irregularity alone. The improvement in visual acuity with a hard contact lens is attributable to irregularity, and the remaining vision impairment with the hard contact lens in place is caused by the opacity. Irregularities are not intrinsically removed by exposure to repeated pulses of the excimer beam. Photons that encounter an irregular surface do not discriminate between elevations and depressions, and tissue that is removed largely replicates the original irregularity (Fig. 3). Strategies have been devised to protect the deeper tissues (valleys) while exposing elevated pathology (peaks) to the ablating photons. Epithelial hyperplasia commonly occurs in areas of corneal depression. Performing transepithelial ablation, where the thickened epithelium shields the valleys from the laser energy, therefore, can take advantage of the natural masking effect of the epithelium. As an example, one strategy for removing surface irregularities of Reis-Bucklers' dystrophy involves ablating through the epithelium that is intermixed with the abnormal substance until the epithelium is largely absent. This absence is determined by the disappearance of the fluorescence that occurs during epithelial ablation. Further ablation then is performed in conjunction with the use of masking fluids. The most commonly used masking fluids are artificial tear substances, which are available in varying viscosity. Kornmehl and coworkers compared artificial tear substances of varying viscosity with saline and a nonfluid control in an experimental model of surface irregularity.64 In a model in which the corneal surface initially was roughened with sandpaper, solutions of moderate viscosity yielded a smoother surface than more viscous artificial tears, and markedly better results than the nonviscous saline solution. All of these fluids outperformed the nonfluid control. A subsequent study found good results with a preparation of 0.25% sodium hyaluronate.65 Figure 4 shows the theoretic concept that a fluid with inadequate viscosity tends to run off, providing insufficient protection for the valleys, whereas a substance that is too viscous drapes over the peaks, protecting the elevations, where ablation is desired. In addition to the effect of viscosity, different solutions may have varying effectiveness in the absorption of 193-nm photons. A thin film of fluid must have adequate absorption to prevent ablation of the underlying tissue. Enhancing absorption at 193 nm may be one step toward the development of a more effective masking fluid for phototherapeutic keratectomy of an irregular surface. A substance such as a gel also could be applied uniformly over the cornea and molded into a new, smooth anterior surface. This gel would be ablated and, as ablation progresses, would expose the peaks while protecting the valleys without the need for repeated artful application of viscous fluids. For this gel to be fully effective, however, the ablation rate of the gel would have to match that of the pathologic tissue. Such a gel has yet to be implemented in clinical practice.66,67  | 
| SUPERFICIAL CORNEAL OPACITIES | 
| In the past, the only options for removal of corneal opacities were lamellar
or penetrating keratoplasty. With the use of the excimer laser, some
superficial opacities can be removed without the need for more invasive
surgery. Figures 5 and 6 show examples of corneal opacities treated with PTK, resulting in marked
improvement in central corneal clarity. Maloney and coworkers reported
the results of 232 patients treated with PTK for corneal visual loss (resulting
from opacities, surface irregularity, or both) in
a prospective, multicenter trial.68 Approximately 45% of patients at each follow-up visit experienced
a two-line improvement of best corrected vision. Other studies have
revealed similar results.69–72  In the evaluation of a visually significant corneal opacity, it is critical to assess its depth. In some conditions, such as Reis-Bucklers' dystrophy, the location of the abnormality is limited to the anterior stroma. In lattice or granular dystrophy, the surgeon must examine the optically significant opacities in the visual zone and assess their depth. Postinflammatory scarring is more difficult to assess because the posterior extent of the opacity typically is ill defined and often is deeper than initially appreciated at slit-lamp examination. However, partial removal of the opacity may be adequate for the patient's visual needs. An optical pachymeter can be a useful adjunct in measuring the depth of the opacity, although the opacity itself may preclude accurate use of this device.73 Visual impairment associated with these disorders often results from a combination of the opacity itself and accompanying corneal irregularity, both of which must be treated to obtain an optimal visual result. In addition, depending on its cause, the opacity may ablate at a different rate than the surrounding stroma. For example, the hyaline material in granular dystrophy ablates more slowly than the surrounding stroma. Therefore, even a surface that was smooth before surgery can become irregular after PTK if careful attention is not given to maintaining a smooth surface during the procedure. It is often effective to debride discrete elevated opacities such as Salzmann nodules or calcium deposits manually with a blade before PTK and then use masking agents to expose the elevated remaining irregularities during the ablation to create or maintain a smooth optical surface.74 Depending on the excimer laser used and the surgeon's preference, ablation usually is performed as a planar disc ablation alone or followed by a hyperopic ablation. These approaches are intended to reduce the frequently induced hyperopic shift. Although the initial ablation may be planar and may retain the original anterior corneal curvature, remodeling after laser treatment with new collagen formation and epithelial hyperplasia at the periphery of the treatment zone may result in net flattening and a hyperopic shift of up to +8.00 D in some cases.57 Amm and Duncker reported that, of 45 patients treated, all refractions remained stable after PTK for recurrent erosions, whereas after treatment for corneal scars, anterior stromal dystrophies, or surface irregularities, 40.6% of patients developed a hyperopic shift, 9% developed a myopic shift, and 40.6% remained stable.75,76 Overall, it was found that deeper ablations were associated with a greater likelihood of hyperopic shift.77 Katsube and coworkers have proposed that the fluid-filled porous nature of the cornea accounts for much of the tendency for the hyperopic shift after PTK.78 In assessing suitability for PTK, the patient's refractive status is a major component. Unless the patient is moderately myopic before surgery, and either has bilateral disease or readily accepts contact lens correction in the fellow eye, the potential for unacceptable anisometropia is high. However, postoperative use of a hard contact lens often is of additional benefit to the patient by optically eliminating residual irregular and regular astigmatism. In addition to the hyperopic shift, deeper ablation, particularly exceeding 50 to 100 μm, usually is accompanied by haze that can become visually significant at its extreme. It is unclear whether intensive steroid administration after laser treatment significantly retards scarring, and to what extent intensive steroid treatment leads to a more pronounced permanent hyperopic shift by retarding wound healing. The brief application of topical mitomycin-C immediately after the ablation also may reduce scar formation.79 Although PTK may reduce the optical and surface disruption of anterior dystrophies, most dystrophies recur naturally, requiring a repeat PTK or transplantation.80 However, Miller and co-workers reported a case of Reis-Bucklers dystrophy that experienced two rapid recurrences after conventional PTK but had 1 year without evidence of recurrence after a third PTK in conjunction with topical application of mitomycin-C. 81 Prolonged use of mitomycin-C has been reported to cause permanent corneal edema, however.82 PTK does not impair the outcome of subsequent penetrating keratoplasty when surgical intervention is needed.83  | 
| TREATMENT OF PRK AND LASIK COMPLICATIONS | |||
| The major interest in the excimer laser has been in the alteration of the
anterior corneal contour for the correction of refractive errors. Improvements
in the techniques and instrumentation for PRK and LASIK have
minimized the risk of significant visual loss associated with these
procedures. However, irregular astigmatism or scarring secondary to the
ablation, the patient's healing response, or flap complications
can still occur. Epithelial hyperplasia and hypoplasia (see Fig. 7) often minimize the irregularities over time, and superficial haze
after PRK often slowly fades. However, in some cases, loss of best corrected
vision, multiplopia, or other optical aberrations can persist
and need to be managed surgically. PTK can be helpful in the management
of these complications. CENTRAL ISLANDS Central islands, usually defined as at least 1.00 D of central steepening in the presence of overall flattening on corneal topography, are often evident after PRK and less commonly after LASIK (Fig. 8). Patients with symptomatic central islands may experience mild loss of best corrected visual acuity, multiplopia, ghost images, halos, glare, and increased astigmatism. The refraction usually is mildly myopic because of the central steep zone. 
 Preventing central island formation requires careful adherence to the manufacturer's recommendations. These include “anti-island” ablation treatments in which extra pulses are delivered to the center of the ablation zone. Some surgeons interrupt the treatment several times to dry the central stroma with a cellulose surgical sponge. The natural history of central islands after PRK is that most resolve or become optically insignificant by 6 months after surgery.2 Although histologic documentation of the healing mechanism of central islands remains unavailable, a common hypothesis is that the central island represents unablated stromal tissue and that the “healing” consists of a thickening of the epithelium in the midperiphery surrounding the island, which gradually envelops and covers the island. After PRK, it is generally prudent to wait at least 6 months in patients with symptomatic central islands before considering retreatment. Some topographically evident central islands do not cause symptoms and should be ignored. When symptoms are present and unchanged after 6 months, and the topography and refractive status are stable, treatment should be considered. After LASIK, in contrast, surface remodeling is minimal. If a symptomatic central island is stable for at least 2 months, relifting the flap and treating the central island in the stromal bed can be considered. Two treatment options for central islands after PRK have been advocated. In the first, the central epithelium is removed with a blade, exposing the central stromal island. The peak of the island is then treated either with a PRK algorithm, whose optical zone is set to the diameter of the central island and whose dioptric correction matches the height of the island, or with a PTK algorithm, again with the diameter set to the width of the island and the number of pulses calculated to create the dioptric flattening needed (approximately 12 pulses per diopter). In either case, the dioptric correction is based on the height of the island as detected on an early topography, usually the 1-month topographic analysis. By 3 to 6 months, the island may appear to be less elevated as a result of partial hyperplasia of the surrounding epithelium; therefore, retreatment will be incomplete. An alternative treatment technique is to retain the epithelium and use a PTK disc ablation equal to the largest diameter of the island. The first area of stroma to be exposed is at the peak of the island if epithelial hyperplasia has occurred around the central island. This can be visualized as the “breakthrough” of the cobalt blue fluorescence that occurs with ablation of the epithelium; fluorescence readily visualized when the room lights and microscope lights are shut off. The ablation through the epithelium progressively exposes the island, whereas the thicker peripheral epithelium shields the surrounding stroma. When the central nonfluorescent stromal island reaches the width of the topographic island as it was seen at 1 month, the island ablation is complete. If the breakthrough of the transepithelial ablation does not occur centrally but rather simultaneously throughout the optical zone, then the surgeon employs PRK or PTK pulses, calculated as described in the preceding paragraph. In either of these techniques, after removal of the tissue island, aggressive topical steroids should be used until stable healing and a stable refractive status are achieved to discourage any central scarring. After LASIK, central islands are treated by lifting the pre-existing flap and ablating the central stroma with a diameter and depth corresponding to the width and height of the island on the topographic map. The authors' preference is to program a myopic correction for the dioptric value of the height of the island on topography. The width is programmed to match the topographic diameter of the island. In the United States, where Food and Drug Administration restrictions may not allow the surgeon to select a small optical zone in myopic corrections, the surgeon using a broad-beam laser can either manually stop the exposure when the ablation is observed to reach the desired size, or, more precisely, apply the number of pulses that correspond to a certain diameter (the laser manufacturer can supply a nomogram interrelating dioptric correction, diameter, and number of pulses). Corneal Haze after PRK After PRK, subepithelial corneal haze usually develops in several weeks, peaks within 1 to 2 months, and then resolves spontaneously during the next 6 to 12 months (Fig. 9). Haze is more prominent in eyes after treating higher degrees of correction, especially when using smaller ablation zones. It is often asymptomatic, being more noticeable to the examiner than the patient. However, sometimes haze can contribute to loss of contrast sensitivity or even loss of best corrected visual acuity. 
 Significant subepithelial haze may improve with increasing the frequency of topical steroids.85 If visually significant haze persists without improvement after 6 to 12 months, surgical intervention may be indicated. Often haze formation is accompanied by regression, so that repeat excimer ablation can simultaneously treat both the regression and the haze. However, retreatment in patients with significant haze yields less predictable results than retreatment of patients with regression in the absence of haze.86 Seiler and associates66 used a combined PTK-PRK approach to treat 21 undercorrected eyes with corneal haze and found scar recurrence in four (19%). Pop and Aras64 found that patients with greater haze had a 40% rate of overcorrection. To limit overcorrection, retreatment should be directed toward decreasing haze to a clinically insignificant level rather than eliminating it. A transepithelial approach often is used in PTK mode with a large ablation diameter. When using the PRK mode, it has been recommended to correct only 50% to 60% of the myopia to avoid overcorrection.64,87 PTK has been successfully combined with amniotic membrane transplants in two cases of severe haze following PRK.88 DECENTERED ABLATIONS Decentered ablations can occur with PRK or LASIK secondary to initial improper positioning or loss of patient fixation with drift of an initially centered treatment (Figs. 10 and 11). During the ablation, the patient's head must be stabilized, typically by the surgeon's hands, and the patient and surgeon must monitor the fixation beams. Decentered ablations are more common after correction of higher degrees of refractive error, probably secondary to the longer fixation time required. They are more likely to be symptomatic with the use of smaller optical zones, because the edge of the ablation is then more likely to be within the pupillary margin. Mild decentrations may be asymptomatic but if greater than 1 mm are often associated with glare, halos, monocular diplopia, and occasionally loss of best corrected visual acuity. 
 Decentrations are difficult to treat; however, some success has been found with using the excimer laser to reapply an ablation.89–92 In one approach, the epithelium is retained over the topographically flattened zone to act as a mask. The epithelium outside the ablation zone is removed either mechanically or with the excimer laser in PTK mode. A ridge at the border of the treatment zone can be ablated as necessary with a small optical zone PTK. A PRK treatment then is performed either centrally or decentered 180 degrees away from the first ablation, with the optical power guided by topography. The already flattened treatment zone is protected by the pre-existing epithelium and, if necessary, other masking agents such as high-viscosity artificial tear fluids or a methylcellulose sponge.93 Pallikaris has described the treatment of decentered ablations without the use of the laser by using a limbal relaxing incision or arcuate keratotomy oriented 180 degrees opposite the decentration.94 Studies are necessary to determine whether these incisions result in better outcomes compared with retreatment with the excimer laser. Decentration usually results in large amount of the high order aberration known as coma. With the clinical introduction of wave front guided optics for primary treatment, the stage is set for development of therapeutic applications. These may be a more precise and effective method for correcting decentration. LASIK FLAP COMPLICATIONS When performing the keratectomy for LASIK, a flap of uniform thickness with a smooth interface is essential. When a thin, incomplete, or buttonholed flap is created, the flap should be replaced without applying the ablation. Unfortunately, irregular astigmatism and scarring with loss of best corrected vision or optical aberrations still can occur and are difficult to treat. PTK potentially can be used to treat the irregular surface in these cases. Kapadia and Wilson reported a case of a buttonholed flap that healed with superficial scarring associated with monocular diplopia and glare.95 Transepithelial PTK followed by PRK for partial treatment of the refractive error was performed with resolution of the optical aberrations. Leu and Hersh used PTK on the underside of the flap and on the stromal bed to resolve a case of recurrent diffuse lamellar keratitis (DLK) that was unresponsive to conventional treatment; ablation of inciting toxins was hypothesized to be the basis for the resolution of the DLK.96 PTK can help resolve recurrent erosions and anterior basement membrane irregularities after LASIK, in a manner similar to the management of primary recurrent erosions as discussed in the preceding.97 FLAP STRIAE Large macrofolds that occur because of flap slippage in the early postoperative period should be treated with immediate repositioning. Long-standing macrofolds, however, are resistant to repositioning and manual smoothing, as are troublesome microstriae that disrupt the anterior corneal optics (Figs. 12 and 13). These complications are described in detail in Chapter 48 of this section. Steinert, Ashrafzadeh, and Hersh reported excellent results with PTK used to treat these irregularities.98 Figure 14 schematically shows the concept of the PTK treatment of striae. The technique, as performed with the VISX broad-beam laser, is to use the PTK mode for planar ablation with the maximal 6.5 mm zone. With the tracker engaged, 200 pulses are applied to the cornea, performing transepithelial ablation. The epithelium acts as a masking agent for the PTK. (The elevated ridges of striae are exposed first, where the epithelium is thinnest, whereas the valleys between the ridges, where the epithelium is hyperplastic, are protected [Fig. 15]). Next, the tracker is turned off. The ring illuminator is turned on, generally revealing a large amount of striae resulting from the removal of the partially masking epithelium. At this point, the surgeon gently debrides any remaining epithelium, stroking with a minimally moist surgical spear sponge or, if needed, a back-tilted surgical blade in the direction away from the hinge. More laser pulses then are applied in short bursts of five or so pulses, with a wipe of masking fluid between each burst. A medium viscosity artificial tear usually is the best choice. The cornea should look slightly moist but not so much fluid that the striae disappear. If the masking fluid bubbles, or the laser pulses sound like a dull thud rather than a snap, too much fluid is present. The PTK procedure stops whenever the cornea looks smoother, but in any case no more than 100 further pulses (300 total) are applied, in order to avoid a major hyperopic shift. Used with this technique, the typical hyperopic shift is about 1 D on the average. In addition, if needed, PRK can be applied after PTK to correct any significant optical error. This will be less accurate than primary PRK due to the variable effect of the PTK, however. Applied with the technique described in the preceding, no optically significant haze was observed in the published series.  | 
| TOPOGRAPHY-BASED ABLATIONS | 
| Despite the PTK techniques available, treatment of irregular astigmatism
remains suboptimal. Fine surface irregularities can be treated using
existing epithelium or masking agents to target the raised areas. However, broad
surface irregularities that often cause irregular astigmatism
after refractive surgery or penetrating keratoplasty are not easily
treated in this fashion. New technology in which topographic data are
incorporated directly into laser software will be helpful for the treatment
of all types of astigmatism. Wiesinger-Jendritza and associates reported a series of 23 eyes treated with LASIK using topography-based ablations for irregular astigmatism after penetrating keratoplasty, trauma, or previous excimer laser surgery.99 They found that 61.9% were partially corrected and 19.4% were fully corrected after the treatment. Improvements in this technique will likely be seen as topographic analysis becomes incorporated into the laser for real-time feedback on topographic changes during the ablation. Vinciguerra and Camesasca also have reported good results with a topography-based custom therapeutic treatment of highly aberrated eyes.100,101 In 52 eyes, the percentage with BSCVA of 20/30 or better increased from a preoperative rate of 59% to 88% at one postoperative year, and BSCVA of 20/15 was reached by 25% of eyes compared to none preoperatively.  | 
| CONCLUSION | 
| Phototherapeutic keratectomy is an important option for patients with painful recurrent erosions. In addition, it offers patients an alternative to lamellar or penetrating keratoplasty for the correction of corneal opacities and surface irregularities. Excimer lasers with real-time corneal topography feedback incorporated into the system will likely provide dramatic improvements in the ability to achieve a smooth corneal surface for patients with irregular astigmatism. This will be invaluable in the treatment of patients with irregular astigmatism after penetrating keratoplasty and previous refractive surgery. In addition, topography-based excimer treatments may allow the surgeon to treat regular astigmatism more precisely and reshape every corneal surface to achieve the optimal corneal contour for best visual results. | 
| REFERENCES | 
1. Burlamacchi P: Laser sources. In Lasers in Biology and Medicine. New York: Plenum, 1980:1 2. Rhodes CK: Excimer Lasers, 2nd ed. New York: Springer-Verlag, 1984 3. Special issue on excimer laser. IEEE J Quantum Electronics OE 15:337, 1979 4. Searles SK, Hart GA: Stimulated emission at 281.8 nm from XeBr. Appl Phys Lett 27:243, 1975 5. McKee T, Nilson JA: Excimer applications. Laser Focus 18:51, 1982 6. Srinivasan R, Leigh WJ: Ablative photodecomposition on poly(ethylene terephthalate) films. J Am Chem Soc 104:6784, 1982 7. Deutsch TF, Geis MW: Self-developing UV photoresist using excimer laser exposure. J Appl Phys 54:7201, 1983 8. Koren G, Yet JT: Emission spectra, surface quality, and mechanism of excimer laser etching of polyamide films. Appl Phys Lett 44:1112, 1984 9. Srinivasan R, Mayne-Banton V: Self-developing photoetching of poly(ethylene terephthalate) films by far-ultraviolet excimer laser radiation. Appl Phys Lett 41:576, 1983 10. Garrison BJ, Srinivasan R: Microscopic model for the ablative photodecomposition of polymers by far-ultraviolet radiation (193 nm). Appl Phys Lett 44:489, 1984 11. Jellinek HH, Srinivasan R: Theory of etching of polymers by far-violet, high-intensity pulsed laser and long-term irradiation. J Phys Chem 88:3048, 1984 12. Melcher RL: Thermal and acoustic techniques for monitoring pulsed laser processing. Springer Series Chem Phys 39:418, 1984 13. Srinivasan R: Kinetics of ablative photodecomposition of organic polymers in the far-ultraviolet (193 nm). J Vac Sci Tech 11:923, 1983 14. Srinivasan R, Braren B: Ablative photodecomposition of polymer films by far-violet (193 nm) laser radiation: Dependence of etch depth on experimental conditions. J Polym Sci Polym Chem Ed 22:2601, 1984 15. Gorodetsky G, Kazyaka TG, Melcher RL, et al: Calorimetric and acoustic study of ultraviolet laser ablation of polymers. Appl Phys Lett 46:828, 1985 16. Slavin W: Stray light in ultraviolet, visible, and near-infrared spectral photometry. Anal Chem 35:561, 1963 17. Beaven GH, Holiday ER: Ultraviolet absorption spectra of proteins and amino acids. Adv Protein Chem 7:319, 1952 18. Loofbourow JR, Gould BS, Sizer IW: Studies on the ultraviolet absorption spectra of collagen. Arch Biochem 22:406, 1949 19. Smith KC: Ultraviolet radiation effects on molecules and cells. In The Science of Photobiology. New York: Plenum, 1985:113 20. Stone AL: Optical rotary dispersion of mucopolysaccharides: III. Ultraviolet circular dichroism and conformational specificity in amide groups. Biopolymers 10:739, 1971 21. O'Brien WJ: Measurement of corneal DNA content. Invest Ophthalmol Vis Sci 18:538, 1979 22. Dougherty AM, Causley GC, Johnson WC: Flow dichroism evidence for tilting of the bases when DNA is in solution. Proc Natl Acad Sci USA 80:2194, 1983 23. Ringvold A: Cornea and ultraviolet radiation. Acta Ophthalmol 58:63, 1980 24. McCann J, Ames BN: Detection of carcinogens as mutagens in the salmonella/microsome test: Assay of 300 chemicals. Discussion. Proc Natl Acad Sci USA 73:950, 1976 25. Nuss RC, Puliafito CA, Dehm EJ: Unscheduled DNA synthesis following excimer laser ablation of the cornea in vivo. Invest Ophthalmol Vis Sci 28:287, 1987 26. Kochevar IE: Cytotoxicity and mutagenicity of excimer laser radiation. Lasers Surg Med 9:440, 1989 27. Muller-Stolzenburg NW, Muller GJ, Buchwald HJ, et al: UV exposure of the lens during 193-nm excimer laser corneal surgery. Arch Ophthalmol 108:915, 1990 28. Trokel SL, Srinivasan R, Braren B: Excimer laser surgery of the cornea. Am J Ophthalmol 96:710, 1983 29. Puliafito CA, Steinert RF, Deutsch TF, et al: Excimer laser ablation of the cornea and lens: Experimental studies. Ophthalmology 92:741, 1985 30. Kerr-Muir MG, Trokel SL, Marshall J, et al: Ultrastructural comparison of conventional surgical and argon fluoride excimer laser keratectomy. Am J Ophthalmol 103:448, 1987 31. Peyman GA, Kuszak JR, Weckstrom K, et al: Effects of XeCL excimer laser on the eyelid and anterior segment structures. Arch Ophthalmol 104:118, 1986 32. Krueger RR, Trokel SL: Quantitation of corneal ablation by ultraviolet laser light. Arch Ophthalmol 103:1741, 1985 33. Krueger RR, Trokel SL, Schubert HD: Interaction of ultraviolet laser light with the cornea. Invest Ophthalmol Vis Sci 26:1455, 1985 34. Puliafito CA, Wong K, Steinert RF: Quantitative and ultrastructural studies of excimer laser ablation of the cornea at 193 and 248 nanometers. Lasers Surg Med 7:155, 1987 35. Fantes EE, Waring GO: Effects of excimer laser radiant exposure on uniformity of ablated corneal surface. Lasers Surg Med 9:533, 1989 36. Berns MW, Liaw L-H, Oliva A, et al: An acute light and electron microscopic study of ultraviolet 193-nm excimer laser corneal incisions. Ophthalmology 95:1422, 1988 37. Dehm EJ, Puliafito CA, Adler CM, et al: Corneal endothelial injury following excimer laser ablation and 193 and 248 nm. Arch Ophthalmol 104:1364, 1986 38. Hanna KD, Pouliquen Y, Waring GO, et al: Corneal stromal wound healing in rabbits after 193-nm excimer laser surface ablation. Arch Ophthalmol 107:895, 1989 39. Zabel R, Tuft S, Marshall J: Excimer laser photorefractive keratectomy: Endothelial morphology following area ablation of the cornea. Invest Ophthalmol Vis Sci 29(suppl):390, 1988 40. Puliafito CA, Stern D, Krueger RR, et al: High-speed photography of the excimer laser ablation of the cornea. Arch Ophthalmol 105:1255, 1987 41. Kahle G, Stadter H, Seiler T, et al: Gas chromatograph/mass spectrometer analysis of excimer and erbium-YAG laser ablated human corneas. Invest Ophthalmol Vis Sci 32(suppl):996, 1991 42. Fagerholm P: Phototherapeutic keratectomy: 12 years of experience. Acta Ophthalmol Scand 81:19, 2003 43. Autrata R, Rehurek J, Vodickova K: Phototherapeutic keratectomy in children: 5-year results. J Cataract Refract Surg 30:1909, 2004 44. Morad Y, Haviv D, Zadok D, et al: Excimer laser phototherapeutic keratectomy for recurrent corneal erosion. J Cataract Refract Surg 24:451, 1998 45. Jain S, Austin DJ: Phototherapeutic keratectomy for treatment of recurrent corneal erosion. J Cataract Refract Surg 25:1610, 1999 46. Lohmann CP, Sachs H, Marshall J, Gabel VP: Excimer laser phototherapeutic keratectomy for recurrent erosions: A clinical study. Ophthal Surg Lasers 27:768, 1996 47. O'Brart DP, Muir MG, Marshall J: Phototherapeutic keratectomy for recurrent corneal erosions. Eye 8:378, 1994 48. Orndahl MJ, Fagerholm PP: Phototherapeutic keratectomy for map-dot-fingerprint corneal dystrophy. Cornea 17:595, 1998 49. Cavanaugh TB, Lind DM, Cutarelli PE, et al: Phototherapeutic keratectomy for recurrent erosion syndrome in anterior basement membrane dystrophy. Ophthalmology 106:971, 1999 50. Dausch D, Landesz M, Klein R, Schroder E: Phototherapeutic keratectomy in recurrent corneal epithelial erosion. Refract Corneal Surg 9:419, 1993 51. Ohman L, Fagerholm P, Tengroth B: Treatment of recurrent corneal erosions with the excimer laser. Acta Ophthalmol 72:461, 1994 52. Bernauer W, De Cock R, Dart JK: Phototherapeutic keratectomy in recurrent corneal erosions refractory to other forms of treatment. Eye 10:561, 1996 53. Forster W, Atzler U, Ratkay I, Busse H: Therapeutic use of the 193-nm excimer laser in corneal pathologies. Graefes Arch Clin Exp Ophthalmol 235:296, 1997 54. Fagerholm P, Fitzsimmons TD, Orndahl M, et al: Phototherapeutic keratectomy: Long-term results in 166 eyes. Refract Corneal Surg 9:S76, 1993 55. Moniz N, Fernandez ST: Efficacy of phototherapeutic keratectomy in various superficial corneal pathologies. J Refract Surg 19:S243, 2003 56. Cavanaugh TB, Lind DM, Cutarelli PE, et al: Phototherapeutic keratectomy for recurrent erosion syndrome in anterior basement membrane dystrophy. Ophthalmology 106:971, 1999 57. Dogru M, Katakami C, Miyashita M, et al: Ocular surface changes after excimer laser phototherapeutic keratectomy. Ophthalmology 107:1144, 2000 58. Sridhar MS, Rapuano CJ, Cosar CB, Cohen EJ, Laibson PR: Phototherapeutic keratectomy versus diamond burr polishing of Bowman's membrane in the treatment of recurrent corneal erosions associated with anterior basement membrane dystrophy. Ophthalmology 109:674, 2002 59. Rosa N, Cennamo G: Phototherapeutic keratectomy for relief of pain in patients with pseudophakic bullous keratopathy. J Refract Surg 18:276, 2002 60. Maini R, Sullivan L, Snibson GR, Taylor HR, Loughnan MS: A comparison of different depth ablations in the treatment of painful bullous keratopathy with phototherapeutic keratectomy. Br J Ophthalmol 85:912, 2001 61. Koksal M, Kargi S, Gurelik G, Akata F: Phototherapeutic keratectomy in Schnyder crystalline corneal dystrophy. Cornea 23:311, 2004 62. Germundsson J, Fagerholm P: Phototherapeutic keratectomy in Salzmann's nodular dystrophy. Acta Ophthalmol Scand 82:148, 2004 63. Stewart OG, Pararajasegaram P, Cazabon J, Morrell AJ: Visual and symptomatic outcome of phototherapeutic keratectomy (PTK) for corneal dystrophies. Eye 16:126, 2002 64. Kornmehl EW, Steinert RF, Puliafito CA: A comparison study of masking fluids for excimer laser phototherapeutic keratectomy. Arch Ophthalmol 109:860, 1991 65. Alio JL, Belda JI, Shalaby AM: Correction of irregular astigmatism with excimer laser assisted sodium hyaluronate. Ophthalmology 108:1246, 2001 66. Katsanevaki VJ, Ginis HS, Naoumidi II, Pallikaris IG: The PALM technique: histologic findings of masked phototherapeutic keratectomy on rabbit corneas. BMC Ophthalmol 3:4, 2003 67. Kremer F, Aronsky M, Bowyer B, Stevens SX: Treatment of corneal surface irregularities using biomask as an adjunct to excimer laser phototherapeutic keratectomy. Cornea 21:28, 2002 68. Maloney RK, Thompson V, Ghiselli G, et al: A prospective multicenter trial of excimer laser phototherapeutic keratectomy for corneal vision loss. The Summit Phototherapeutic Keratectomy Study Group. Am J Ophthalmol 122:149, 1996 69. Sher NA, Bowers RA, Zabel RW, et al: Clinical use of the 193-nm excimer laser in the treatment of corneal scars. Arch Ophthalmol 109:491, 1991 70. Stark WJ, Chamon W, Kamp MT, et al: Clinical follow-up of 193-nm ArF excimer laser photokeratectomy. Ophthalmology 99:805, 1992 71. Campos M, Nielsen S, Szerenyi K, et al: Clinical follow-up of phototherapeutic keratectomy for treatment of corneal opacities. Am J Ophthalmol 115:433, 1993 72. Rapuano CJ: Excimer laser phototherapeutic keratectomy: Long-term results and practical considerations. Cornea 16:151, 1997 73. Stark WJ, Gilbert ML, Gottsch GD, Munnerlyn C: Optical pachymetry in the measurement of anterior corneal disease: An evaluative tool for phototherapeutic keratectomy. Arch Ophthalmol 108:12, 1990 74. Stewart OG, Morrell AJ: Management of band keratopathy with excimer laser phototherapeutic keratectomy: Visual, refractive, and symptomatic outcome. Eye 17:233, 2003 75. Amm M, Duncker GI: Refractive changes after phototherapeutic keratectomy. J Cataract Refract Surg 23:839, 1997 76. McGhee CNJ, Bryce IG: Natural history of central topographic islands following excimer laser photorefractive keratectomy. J Cataract Refract Surg 22:1151, 1996 77. Rapuano CJ: Excimer laser phototherapeutic keratectomy in eyes with anterior corneal dystrophies: preoperative and postoperative ultrasound biomicroscopic examination and short-term clinical outcomes with and without antihyperopia treatment. Trans Am Ophthalmol Soc 101:371, 2003 78. Katsube N, Wang R, Okuma E, Roberts C: Biomechanical response of the cornea to phototherapeutic keratectomy when treated as a fluid-filled porous material. J Refract Surg 18:S593, 2002 79. Porges Y, Ben-Haim O, Hirsh A, Levinger S: Phototherapeutic keratectomy with mitomycin C for corneal haze following photorefractive keratectomy for myopia. J Refract Surg 19:40, 2003 80. Dinh R, Rapuano CJ, Cohen EJ, Laibson PR: Recurrence of corneal dystrophy after excimer laser phototherapeutic keratectomy. Ophthalmology 106:1490, 1999 81. Miller A, Solomon R, Bloom A, et al: Prevention of recurrent Reis-Bucklers dystrophy following excimer laser phototherapeutic keratectomy with topical mitomycin C. Cornea 23:732, 2004 82. Pfister RR: Permanent corneal edema resulting from the treatment of PTK corneal haze with mitomycin: A case report. Cornea 23:744, 2004 83. Szentmary N, Langenbucher A, Hafner A, Seitz B: Impact of phototherapeutic keratectomy on the outcome of subsequent penetrating keratoplasty in patients with stromal corneal dystrophies. Am J Ophthalmol 137:301, 2004 84. Abbas UL, Hersh PS: Natural history of corneal topography after excimer laser photorefractive keratectomy. Ophthalmology 108(5):953, 2001 85. Marques EF, Leite EB, Cunha-Vaz JG: Corticosteroids for reversal of myopic regression after photorefractive keratectomy. J Refract Surg 11:S302, 1995 86. Pop M, Aras M: Photorefractive keratectomy retreatments for regression: One-year follow-up. Ophthalmology 103:96, 1979 87. Gartry DS, Larkin DF, Hill AR, et al: Retreatment for significant regression after excimer laser photorefractive keratectomy: A prospective, randomized, masked trial. Ophthalmology 105:131, 1998 88. Lee HK, Kim JK, Kim EK, Kim GO, Lee IS: Phototherapeutic keratectomy with amniotic membrane for severe subepithelial fibrosis following excimer laser refractive surgery. J Cataract Refract Surg 29:1430, 2003 89. Seiler T, Schmidt-Petersen H, Wollensak J: Complications after myopic photorefractive keratectomy primarily with the Summit excimer laser. In Corneal Laser Surgery. St. Louis: Mosby, 1995:131 90. Alkara N, Genth U, Seiler T: Diametral ablation: A technique to manage decentered photorefractive keratectomy for myopia. J Refract Surg 15:436, 1999 91. Lim-Bon-Siong R, Williams JM, Steinert RS, Pepose JS: Retreatment of decentered excimer photorefractive keratectomy ablations. Am J Ophthalmol 123:122, 1997 92. Talamo JH, Wagoner MD, Lee SY: Management of ablation decentration following excimer laser photorefractive keratectomy. Arch Ophthalmol 113:706, 1995 93. Rachid MD, Yoo SH, Azar DT: Phototherapeutic keratectomy for decentration and central islands after photorefractive keratectomy. Ophthalmology 108:545, 2001 94. Pallikaris I, Siganos D: LASIK complications management. In The Excimer Manual. Boston: Little, Brown, 1997:227 95. Kapadia MS, Wilson SE: Transepithelial photorefractive keratectomy for treatment of thin flaps or caps after complicated laser in situ keratomileusis. Am J Ophthalmol 126:827, 1998 96. Leu G, Hersh P: Phototherapeutic keratectomy for the treatment of diffuse lamellar keratitis. J Cataract Refract Surg 28:1471, 2002 97. Rojas MC, Manche EE: Phototherapeutic keratectomy for anterior basement membrane dystrophy after laser in situ keratomileusis. Arch Ophthalmol 120:722, 2002 98. Steinert RF, Ashrafzadeh A, Hersh PS: Results of phototherapeutic keratectomy in the management of flap striae after LASIK. Ophthalmology 111:740, 2004 99. Wiesinger-Jendritza B, Knorz MC, Hugger P, Liermann A: Laser in situ keratomileusis assisted by corneal topography. J Cataract Refract Surg 24:166, 1998 100. Vinciguerra P, Camesasca FI: Custom phototherapeutic keratectomy with intraoperative topography. J Refract Surg 20:S555, 2004 101. Vinciguerra P, Camesasca FI: One year follow-up of custom phototherapeutic keratectomy. J Refract Surg 20:S705, 2004  |